Viewing and evaluating data

The Data Acquisition Module (DAqM) and Data Retention Module (DRM) are pass-through devices that do not support data logging by themselves. Although the DRM can store up to 32 MB (about 1 month) of biomet data on its internal memory, those records are sent to the SmartFlux 2 or 3 System, where they are stored on the USB storage device on the connected SmartFlux System. The raw biomet data are logged in zipped eddy covariance data files (.ghg file extension) that also include the high-speed gas and wind data. A summary file of biomet data (1-minute records) is stored on the SmartFlux System each day. In the event that the DRM disables power to the SmartFlux System, a backup file of biomet data is saved on the DRM and is transferred to the SmartFlux System USB drive when it is powered back on. Biomet data are used by EddyPro Software on the SmartFlux System and summarized in the EddyPro output files. Long-term data can be evaluated in Tovi Software.

The relationship between variable label and sensor position

Each program in the Biomet Library includes a set of Constants and Variables. Constants represent values that do not change over time, such as a sensor calibration multiplier. Variables represent recorded values that may change over time, such as air temperature. Variable names present a simple way to organize and differentiate measurements from different sensors on the basis of position and replicate number. If you have multiple soil temperature sensors at a station, or multiple eddy covariance stations at a site, variable names will keep them organized.

Anatomy of variable labels

Variables are labeled according to the convention described below. Labels include a alpha-numeric code that identifies the parameter, horizontal position, vertical position, and replicate number (see Figure 6‑1). If you use a custom program, variables are logged with the label you specify in the program. Use the conventions described here to ensure that EddyPro will process your data and to have a simple way to keep track of which sensor is which.

Figure 6‑1. Variable labels indicate the parameter, position, and replicate at a position.
Table 6‑1. Variables that are included in biomet programs. Some variables are only loaded in specific programs.
Label Variable Units
DAQM_V_MAIN_1_1_1 Data Acquisition Module Voltage V
DAQM_T_1_1_1 Data Acquisition Module Temp. °C
SWC_#_#_# Soil Water Content m3/m3
TS_#_#_# Soil Temperature °C
PPFD_#_#_# Photosynthetic Photon Flux Density umol/m2/s
RG_#_#_# Global Radiation W/m2
TA_#_#_# Air Temperature °C
RH_#_#_# Relative Humidity %
P_RAIN_#_#_# Precipitation (rain) mm
RN_#_#_# Net Radiation W/m2
SWIN_#_#_# Shortwave Incoming Radiation W/m2
SWOUT_#_#_# Shortwave Outgoing Radiation W/m2
LWIN_#_#_# Longwave Incoming Radiation W/m2
LWOUT_#_#_# Longwave Outgoing Radiation W/m2
TNR_C_#_#_# CNR4 or NR01 Net Radiometer Temperature °C
TNR_K_#_#_# CNR4 or NR01 Net Radiometer Temperature K
ALB_#_#_# Albedo none
SHF_#_#_# Soil Heat Flux W/m2
SHFSENS_#_#_# Soil Heat Flux Sensitivity uV/W/m2
SHFHEATER_#_#_# Soil Heat Flux Plate Heater V

Sensor position and replicate number

The position and replicate number of each sensor is indicated by the variable label. Figure 6‑2 shows an example with biomet sensors and variable labels. Figure 6‑3 shows an example with two towers and additional air temperature sensors.

Figure 6‑2. Variables are identified by a label. The letters specify which variable is measured. The subsequent numbers specify the horizontal position, vertical position, and replicate number if locations are identical.
Figure 6‑3. If there are multiple sensors at the same horizontal position but different vertical positions, this is indicated by the first digit (SHF_1_1_1 and SHF_2_1_1). Replicates at the different horizontal positions but the same vertical postion are indicated by the second digit (TA_1_1_1 and TA_1_2_1). Replicates at the same horizontal and vertical location are indicated by the third digit (TA_1_1_1 and TA_1_1_2 on Tower 1 and TA_2_1_1 and TA_2_1_2 on Tower 2). In this image, the two towers have replicates of air temperature measurements on both vertical axes and horizontal axes, as well as redundant air temperature measurements at some positions.

Although EddyPro can "blindly" summarize any variable, including variables that are not listed here, it is unable to use the variable in flux computations (e.g., air temperature) unless the variable is identified according to the naming convention described in this section.

Viewing live data from the instruments

Measured values can be viewed in the gas analyzer PC application software (known as the GHG software for the LI-7500A/RS/DS and LI-7200/RS). Data are refreshed every 60 seconds. These data provide a quick check to see if sensors are providing plausible values and responding as expected.

In the gas analyzer software, click Biomet. Under the Data tab, you'll see a list of sensors, units, and data values.

Viewing data in FluxSuite

FluxSuite Software (fluxsuite.com) is a cloud-based application that provides near real-time access to site status information, processed flux results, biomet data, and images taken with a PhenoCam. Contact LI-COR or your distributor for more information.

FluxSuite Software

Copying data files from the SmartFlux System

Files can be managed using the Download utility that is built into the gas analyzer PC software or a Secure File Transfer Protocol (SFTP) application.

Using the Download utility

The Download utility allows you download and delete files that are stored on the USB device in the SmartFlux System. You can access the files remotely, so long as you can connect to your instrument over a network. The utility is a component of the gas analyzer software. Click the Download button to launch it.

The application should automatically connect to the instrument, but if it doesn't, click Browse to view a list of instruments on the network. Select the instrument, click OK, enter the password (licor) and click Verify Connection. You may need to set the port if you are connecting with a cellular or satellite gateway. If your instrument is not listed, type the IP address into the IP Address field. If you are using a RV50X, set the Port to the same number that you have opened in the RV50X software.

USB Drive Information displays the size of the USB drive, available memory, number of each type of file.

Transfer Options are used to configure settings and other features:

  • Transfer Now will allow you to use WinSCP to transfer files. WinSCP is a free file transfer application that can be installed as part of the win-GHG application. Click Launch WinSCP to open the application.
  • Destination: This is where the transferred files will be saved. Ideally, this will be a directory on your computer. If you are transferring files to a server or cloud-based service such as a Dropbox, this is configured in the SmartFlux section of the application.
  • Schedule Transfer: This section allow you to configure automatic file downloads to a computer.
  • When to transfer: You can Transfer files daily at a scheduled time.
  • You can close the application to your system tray (close it, but choose No) so it continues running in the background. Also be sure your computer does not go to sleep before the scheduled transfer. You may need to re-start the automatic download service if your computer is shut down and re-started.
  • Files to transfer: Select any or all file types to transfer data files (you must select a file type before files are displayed). When you check a box, a corresponding tab is added to the box on the right. Here you can choose to download specific files or all of them.
  • After configuring the transfer and selecting files, click Start Timer and allow the application to run in the background so it can download files as scheduled.

Tips for the Scheduled Download utility: 

  • If the downloader is already running, open it from your system "tray" rather than starting another instance of it.
  • Set the download to begin at 00:15 (12:15 am) if you are using the SmartFlux System. That way all the logged files for the previous day will be processed before the download begins.
  • Close the application to your system tray to keep it running in the background. Be sure the computer that runs the application is on and not in power saver mode.

Using an SFTP application

You can also use third-party file transfer (SFTP) applications to transfer data, such as WinSCP or another application. In this case, connect using the instrument IP address and the following settings:

  • Port: 22 or 2200 if you have forwarded port 22 to port 2200.
  • Username: licor
  • Password: licor

You will see all of the files in the file system. The files of interest are in a data folder.

Viewing logged data

To view raw data, copy raw data files from the SmartFlux System in one of two ways: connect the EC system to your computer or network and download the data using the gas analyzer PC application software, or eject the USB storage device from the SmartFlux System and connect it to your computer to read the data.

In the LI-7x00 File Viewer

The LI-7x00 File Viewer application (v1.0.4) is included with the gas analyzer software. If you didn't install with your gas analyzer application software, you can download it from www.licor.com/env/support. To view data:

  1. Launch the LI-7x00 File Viewer application and open the .ghg file.
  2. Files are opened by dragging and dropping into the File Viewer interface or by clicking the plus button (+) and selecting the files. You can load one or many files. They'll be in the order of the time stamps. When a file is open, you'll see a list of variables under the Variables tab on the right.
  3. View one or more variables.
  4. Double-click any variable to see a chart of the time-series. Drag any other variable onto the chart to view multiple variables. Use the zoom controls to examine the data in more detail.
Figure 6‑4. Three replicates of soil heat flux measurements covering a 24-hour period. The File Viewer can load and plot nearly any variable from your dataset.

The file viewer also provides options to export the dataset, which you can use at your discretion.

In a text editor

If you want to see text data, you open the daily text file of biomet data (logged only on the SmartFlux System USB drive), or your can extract the biomet data files from the compressed eddy covariance data files (.ghg files saved on both the SmartFlux System USB drive and the LI-7550 USB drive, if applicable).

  1. Copy the file of interest.
  2. Do not alter the original data. Copy the files before modifying them so that you are not altering the original data. If you alter the original data, EddyPro may be unable to process it.
  3. Open the file directly in 7zip compression application.
  4. If you don't use 7zip, change the file extension from .ghg to .zip.
  5. For example, the file name 2018-05-05T003000_AIU-0288.ghg should be changed to 2018-05-05T003000_AIU-0288.zip.
  6. Extract the contents of the zipped file.
  7. Biomet data are stored in a file with a name like 2018-05-05T003000_AIU-0288-biomet.data. The file contains a file header, data header, and data. The file header describes the system that collected the data. The data header identifies the variable and units. The data (1-second records) are in tab-delimited columns. A second file, with a name like 2018-05-05T003000_AIU-02880-biomet.metadata, includes information (metadata) about the biomet data. You can open either of these files in a text editor or spreadsheet.

Although you can read the raw data and use it, we don't recommend working with the .ghg files outside of EddyPro Software. EddyPro will summarize and average the biomet data.

Figure 6‑5. Raw biomet data are logged in text files.

Biomet data logged while on backup battery

For systems equipped with a DRM and backup battery, when the system switches to backup battery power, biomet data will be logged in a seperate text file while the gas analyzer, SmartFlux System, sonic anemometer, and accessories are powered off. These files are temporarily stored on the DRM and copied to the SmartFlux System USB drive when power is restored. This file will have the same time stamp as a typical biomet data file, representing the year, month, day, and time, but be appended with the word backup. It will be compressed into a .zip file. The format is similar to a biomet data file, but with a few minor changes.

Viewing processed data from the EddyPro output

Logged biomet data are processed by EddyPro, and in a typical eddy covariance system, these data will be recorded as 30 minute averages with flux data. The EddyPro output files include a file called something like eddypro_Demo_biomet_2018-05-05T143614_adv.csv. Data are stored as comma separated values (.csv) in a text file that can be opened in a spreadsheet or text editor. The header displays the variable label and position code (see Support: Biomet Data Acquisition System and Blueprint Utility), as well as the variable units.

Figure 6‑6. EddyPro results can be opened in a spreadsheet. Results are 30-minute averages.

Evaluating long-term data in Tovi® Software

Tovi Software provides tools to evaluate and process the fluxes and biomet data that have been processed by EddyPro. For example, you can average redundant measurements, merge variables and gap fill with data from your site, gap fill with data from nearby weather stations, evaluate the quality of data, and use the biomet data to interpret flux results. Tovi is designed to evaluate lengthy time-series eddy covariance data sets.

Figure 6‑7. A one-month data set showing three replicate measurements of soil temperature (K). Tovi Software can be used to evaluate, merge, gap fill, and average replicate measurements of a parameter.