Using the Gill HS and R3 Anemometers in Eddy Covariance Systems

Printable PDF: Using the Gill HS and R3 Anemometers in Eddy Covariance Systems

(EC_InstallGuide_Gill_HS_R3-17988)

Download this content as a pdf that can be saved to your computer or printed.

This document describes how to use the Gill HS-50, HS-100, and R3-50 sonic anemometers in LI-COR eddy covariance systems, using the HS-50 as an example. The topics covered are installing the anemometer at your site, connecting the anemometer to the SmartFlux 2 or 3 System, and configuring the software for the eddy covariance site.

Important: Be sure to read and understand the documentation provided with the sonic anemometer.

Table 1‑1. Power and data cables to connect Gill HS or R3 anemometers with the SmartFlux System.
HS-50 Cable Type Length Part Number
Data and Power 5 m 7900-445-5
Data and Power 25 m 7900-445-25
Data and Power 50 m 7900-445-50
USB to RS-422 1.8 m 392-16348

Installing the anemometer and analyzer

The HS-50 anemometer can be mounted to a horizontal cross-arm with a crossover fittings and post hardware. The R3-50 mounts with the mounting apparatus (7900-342).

Figure 1‑1. The HS-50 mounts to a vertical riser bar.

Orientation, horizontal offsets, and vertical offset

Mount the anemometer oriented toward magnetic north, then rotate it so that it faces the prevailing wind. Measure and record the north offset and height.

  • North offset (°): The degrees of rotation from magnetic north.
  • Height (m): The distance between the ground and the center of the anemometer sample volume.
Figure 1‑2. Overhead view of the HS-50. The north offset is the degrees of rotation from magnetic north.

Horizontal separation distances (Figure 1‑3) are measured in a Cartesian coordinate system, which EddyPro® uses to determine the distance between the gas sample and the anemometer. The center of the anemometer is the center (0, 0) of the coordinate system. Distances from the anemometer are provided along the north-south and east-west axes. The sensor head or air inlet should be 10-20 cm from the anemometer in the horizontal plane.

Figure 1‑3. Horizontal separation distances.

For deployments that are close to a plant canopy, the center of the analyzer and sonic anemometer should be at equal heights, as shown in Figure 1‑4. If mounted over 15 m above ground level, horizontal and vertical sensor separation can be as large as 30-50 cm. Measure and record the horizontal and vertical separation distances.

Figure 1‑4. Measure the vertical offset between the analyzer sample volume or inlet and the anemometer midline.
  • Northward Separation (cm): North (positive values) or south (negative values) distance between the sample path or intake and the anemometer.
  • Eastward Separation (cm): East (positive values) or west (negative values) distance between the sample path or intake and the reference anemometer.
  • Vertical Separation (cm): Vertical distance between the sample path or inlet and the reference anemometer. Negative if the center of the analyzer or the inlet is below the center of the reference anemometer sample volume; positive if the inlet is above.

Connecting to the SmartFlux 2 or 3 System

The anemometer is automatically configured by the SmartFlux System when connected and powered on. The anemometer cable connects to port 1, 2, or 3 on the SmartFlux System (port 1 shown). Port 1 accepts 11-pin and 9-pin terminal strips. Ports 2 and 3 accept 9-pin terminal strips.

Important: Disconnect the power supply to the SmartFlux System before connecting anemometer cables.

Table 1‑2. Wiring for the HS-50, HS-100, and R3-50 data and power cable. The three black wires are bundled with their counterpart in the cable.
SmartFlux Terminal SmartFlux Label Wire Color
9 RS-422/485 RX- Black (bundled with white)
8 RS-422/485 RX+ White
7 RS-422/485 TX- Black (bundled with brown)
6 RS-422/485 TX+ Brown
5 Signal Return Red
2 Power Out (-) Black (bundled with yellow)
1 Power Out (+) Yellow
Figure 1‑5. Wiring the HS-50 anemometer to the SmartFlux System. The R3 is similar, but the cable connects directly to the anemometer, rather than the electronics control unit.
Figure 1‑6. The HS-50 and SmartFlux System are joined with the gas analyzer using a network switch. The R3 is similar, but the cable connects directly to the anemometer.

Important: At least 13 VDC must be supplied when using a 50 meter power cable because of voltage drops.

Configuring the eddy covariance system

Several things need to be configured in software before your system will log valid eddy covariance data sets. These are described below.

Enter sonic anemometer information

In the gas analyzer PC software, click Site Setup > Sonic Anemometer tab, and select Anemometer connected to a digital port on the SmartFlux 2 or 3 System.

  • Port: The port on the SmartFlux system that the anemometer is connected to (1 in this example).
  • Manufacturer: Gill .
  • Model: HS-50, HS-100, or R3-50.
  • Serial Number: Optional, the anemometer serial number.
  • Software Version: Optional, the anemometer software version.
  • Wind Data Format: U, V, W is the only option.
  • North Offset: Degrees clockwise of magnetic North. Measured at the site (see Figure 1‑2).
  • Height: Distance of the anemometer sample volume above the ground. Measured at the site.

Enter gas analyzer information

The following information is required:

Additional system settings

Although not related to the sonic anemometer, configure the following settings to finish configuring the eddy covariance system. Under Site Setup > USB Log File > USB Log File Options, set the following:

  • Log Rate (Samples per Second): 10.0 (or higher).
  • File Duration: 30 minutes.

Under Site Setup > Site Description, set the following:

  • Site Name: A name of your choice.
  • Station Name: A name of your choice.
  • Canopy Height (m): The distance between ground level and the top of the plant canopy.
  • Check Use SmartFlux GPS coordinates.

At this point, you've configured the minimum required settings to log complete eddy covariance data sets with an LI-7500A/RS/DS. There are additional settings to configure if you are using an LI-7200/RS, a biomet system, or an LI-7700, or if you want to process data using EddyPro Advanced settings. These are described in corresponding product documentation.