Equation summary

The LI-6800 measures and computes the following parameters related to leaf-level gas exchange measurements. Additional details are given in Theory and equation summary

Light

The photosynthetic photon flux density (PPFD) readings Qin and Qout are computed from

9‑19

9‑20

where Sqin and Sqout are the calibration factors, and Iqin and Iqout are the raw counts from the A/D converter.

Leaf temperature

Leaf temperatures Tl1, Tl2 (°C) for the two thermocouples are computed from signals Vl1 and Vl2 (mV) by

9‑21

9‑22

where Tj1 and Tj2 are junction temperatures (°C).

Pressure

Atmospheric pressure Pa (kPa) is reported directly from the on-board sensor.

Chamber over-pressure ΔPc (kPa) is computed from

9‑23

where Vp is the signal from the differential pressure sensor (in V, but taken to be kPa) Vpo is the calibration zero for the sensor, G is fan speed (rpm), F is flow rate, and pca and pcb are empirical coefficients that depend on chamber type (Table 9‑1).

Table 9‑1. Values for pressure correction coefficients for each chamber type.
Chamber pca pcb
6800-01 Fluorometer 0 0
6800-01A Fluorometer 4.3321E-10 1.0215E-5
6800-12 3x3 0 0
6800-12A 3x3 2.6520E-10 1.3064E-5
6800-13 6x6 4.3998E-11 1.4792E-5
6800-17 Small Plant 0 0
6800-19 Custom 0 0

Flow

The flow F (µmol s-1) to the leaf chamber is computed from

9‑24

where af1...af7 are factory calibration constants, Vf is flow meter signal (V), Vfo is calibration zero, and Tk is IRGA block temperature.

Reference and sample IRGA cell flows Fr and Fs (µmol s-1) are computed from

9‑25

9‑26

where Vfr and Vfs are cell exit flow sensor voltage, Vfro and Vfso are the offset value of that voltage, and ar1...ar4 and as1...as4 are factory calibration coefficients.

CO2

Reference and sample CO2 concentrations Cr and Cs are given by

9‑27

where Mc() is the CO2 match correction function:

9‑28

9‑29

9‑30

where αcr and αcs are absorptances, Pa is atmospheric pressure, Tr and Ts are the cell inlet temperatures, fc is a 5th order polynomial with coefficients (br1...br5 or bs1...bs5) determined at the factory, Scro; Scso and Scr1; Scs1 are the span1 and span2 user calibration settings, Wa and Wb are water concentrations, Xo is the oxygen concentration in percent. Ѱc is the band broadening function for the effect of water and oxygen on CO2, given by

9‑31

where Bwc and Boc are the band broadening coefficients for H2O and O2 respectively on CO2.

Reference and sample absorptances αcr and αcs are corrected for zero drift with temperature, span drift with temperature, and cross sensitivity with water.

9‑32

9‑33

where cr1...cr3 and cs1...cs3 are empirical coefficients for CO2 absorptance span drift determined during calibration, Icr, Icro, Ics, and Icso are the raw IRGA detector absorbing and non-absorbing readings for CO2, and Iwr, Iwro, Iws and Iwso are those for water. Xwcr and Xwcs are empirical cross sensitivity coefficients for water on CO2 determined during calibration, Zcro and Zcso are the current user CO2 zero factors, and Zcr and Zcs are the factors for CO2 zero drift with temperature, determined during calibration.

Dry mole fractions Crd and Csd are computed from

9‑34

H2O

Reference and sample H2O concentrations Hr and Hs are given by

9‑35

where Mw() is the H2O match correction function:

9‑36

9‑37

9‑38

where αwr and αws are the reference and sample absorptances, fw is a 3rd order polynomial with coefficients (dr1...dr3 and ds1...ds3) determined at the factory, Swro, Swr1 and Swso, Sws1 are the H2O span1 and span2 user calibration settings for reference and sample, Ѱo is the band broadening function for the effect of oxygen on H2O, given by

9‑39

where Bow1 and Bow2 are empirical coefficients.

Absorptances αwr and αws are corrected for zero drift with temperature, span drift with temperature, and cross sensitivity with water.

9‑40

9‑41

where wr1...wr3 and ws1...ws3 are empirical coefficients for H2O absorptance span drift determined during calibration. Xcwr, Xcws are empirical cross sensitivity coefficients for CO2 on H2O determined during calibration, Zwro and Zwso are the current user H2O zero factors, and Zwr and Zws are the factors for H2O zero drift with temperature, determined during calibration.

The vapor pressure (kPa) of the air in the reference and sample cells er and es is given by

9‑42

9‑43

Reference and sample cell dew point temperatures Tdr and Tds are given by

9‑44

9‑45

Humidity indicators

The leaf chamber vapor pressure ec (kPa) is given by

9‑46

The saturation vapor pressure esc in the leaf chamber is a function of chamber air temperature Ta:

9‑47

where es(T) is the saturation vapor pressure function:

9‑48

The relative humidity hc (%) in the leaf chamber is given by

9‑49

The vapor pressure deficit of the leaf eΔl is computed from

9‑50

where Tl is leaf temperature ().

Transpiration

The mass balance of water vapor in an open system is given by

9‑51

where s is leaf area (m2), E is transpiration rate (mol H2O m-2 s-1), ue and uo are incoming and outgoing flow rates (mol s-1) from the chamber, and we and wo are incoming and outgoing water mole factions (mol H2O (mol air)-1). Since

9‑52

we can write

9‑53

which rearranges to

9‑54

The relationships between the terms in equations 9‑51 through 9‑54 and what the LI-6800 measures are

9‑55

where F is air flow rate (µmol s-1), Ws and Wr are sample and reference water mole fractions (mmol H2O (mol air)-1), and S is leaf area (cm2). The equation that the LI-6850 uses for transpiration is thus

9‑56

Stomatal conductance

The total conductance gtw of the leaf to water vapor is

9‑57

where Wl is the molar concentration of the water vapor within the leaf (mmol H2O (mol air)-1), which is computed from the leaf temperature Tl (°C) and the total atmospheric pressure in the leaf chamber

9‑58

We assume that the total resistance for the upper ru or lower rl surface of a leaf is the sum of the stomatal resistance rs and boundary layer resistance rb of that surface

9‑59

9‑60

and that the upper and lower boundary layer resistances are the same

9‑61

and we define K to be the ratio of stomatal resistances of the two sides

9‑62

Leaf stomatal resistance rs is given by

9‑63

9‑64

9‑65

9‑66

Total conductance g (the inverse of the total resistance r) can thus be written

9‑67

9‑68

9‑69

For water vapor, the total conductance gtw is related to stomatal conductance gsw and one sided boundary layer conductance gbw by

9‑70

Solving equation 9‑68 for gsw yields

9‑71

Note that although we defined K to be a particular ratio of upper to lower stomata resistances (equation 9‑62), since K appears as , we get the same mathematical result if we had defined it the other way. In other words, K is equivalent to 1/K. Note also that the LI-6400 does not use equation 9‑71, but rather a simplified approximation:

9‑72

Boundary layer

The one sided boundary layer conductance to water vapor gbw for a broadleaf is a function of fan speed G (rpm) and leaf area S (cm2).

9‑73

where f is

9‑74

and s is forced to be

9‑75

The empirical coefficients co...c4, reference pressure Po, and leaf area limits Smin and Smax depend on chamber type.

Table 9‑2. Values for boundary layer conductance for chamber types.
Chamber co c1 c2 c3 c4 Po Smin Smax
6800-01 Flr 0.250 0.35860 -4.01816E-3 0.00451074 -0.0044762 96.9 1 6
6800-01A 6 cm2 0.578 0.5229739 3.740252E-3 -6.197961E-2 -5.608586E-3 96.9 1 6
6800-01A 2 cm2 0.572 0.3872742 -1.870584E-2 0.0 -7.37389E-3 96.9 1 2
6800-12 3x3 0.500 0.44869569 1.9000035E-3 -4.26088781E-2 -3.456516E-3 96.7 2 9
6800-12A 9 cm2 0.579 0.3210639 -1.109987E-3 5.106816E-3 -3.283688E-3 96.7 2 9
6800-12A 6 cm2FB 0.345 0.552336 -4.7985e-3 0.0 -7.3557e-3 96.7 1 6
6800-12A 6 cm2SS 0.418 0.5145466 -2.5106E-3 0.0 -8.1206E-3 96.7 1 6
6800-12A 3 cm2FB 0.188 0.5795409 -1.15295E-2 0.0 -9.7259E-3 96.7 1 3
6800-12A 3 cm2SS 0.141 0.5263354 -1.27376E-2 0.0 -1.10157E-2 96.7 1 3
6800-13 6x6 0.430 0.267827 -1.164018E-4 2.248202E-3 -5.109462E-3 96.8 6 36

Net assimilation

The mass balance of CO2 in an open system is given by

9‑76

where a is assimilation rate (mol CO2 m-1 s-1), ce and co are entering and outgoing mole fractions (mol CO2 (mol air)-1). Using equation 9‑52, we can write

9‑77

which rearranges to

9‑78

To write equation 9‑78 in terms of what the LI-6800 measures, we use equations 9‑55 and

9‑79

where Cr and Cs are sample and reference CO2 concentrations (µmol mol-1), and A is the net assimilation by the leaf (µmol m-2 s-1). Substitution yields

9‑80

9‑81

9‑82

Intercellular CO2

The intercellular CO2 concentration Ci (µmol CO2 (mol air)-1) is given by

9‑83

where gtc is the total conductance to CO2. From equation 9‑69, we can write

9‑84

where 1.6 is the ratio of the diffusivities of CO2 and water in air, and 1.37 is the same ratio for the boundary layer. This is another departure from the LI-6400, which does not use equation 9‑84, but a simplified approximation

9‑85

Energy balance

The LI-6800 provides several potential sources for leaf temperature: it can be directly measured with some combination of the two leaf thermocouples (Tl1 and Tl2), or computed indirectly from a leaf energy balance (Teb). The user can specify what combination to use via three weighting factors fT1, fT2, and fTeb:

9‑86

The energy balance temperature Teb assumes that the energy balance of a leaf in the chamber has three components: net radiation R (W m-2), sensible heat flux H (W m-2), and latent heat flux L (W m-2), and that they all sum to zero:

9‑87

We consider two components of net radiation: short wave (visible and near IR) and thermal.

9‑88

where Rabs is absorbed short wave, and Rnt is net thermal. The absorbed short wave radiation is computed by

9‑89

where Qabs is the absorbed irradiance by the leaf (µmol m-2 s-1), and k is the conversion factor for transforming (µmol m-2 s-1) to (W m-2), based on the spectral characteristics of the light source.

The net thermal balance is based on the leaf temperature Tl and the surrounding chamber wall temperature Tw, so the total radiation balance R can be written as

9‑90

where ϵ is the thermal emissivity of the leaf (usually assumed to be 0.95), and σ is the Stefan-Boltzmann constant (5.67 W m-2 K-1). The 2 in equation 9‑90 accounts for both sides of the leaf. Wall temperature Tw is not measured, but depends on a user-specified offset ΔTw from chamber air temperature.

9‑91

The latent heat flux L is the transpiration rate E converted to W m-2.

9‑92

The sensible heat flux H is a function of the leaf - chamber air temperature difference TlTa, the specific heat capacity of the air cp (29.3 J mol-1 K-1) and the one sided boundary layer conductance for heat transfer of the leaf gbh, which is 0.92 times the boundary layer conductance for water vapor gbw.

9‑93

9‑94

Equation 9‑87 becomes:

9‑95

If we let ΔT = TlTa, and note for small ΔT

9‑96

Substituting equation 9‑96 into 9‑95 and solving for ΔT yields

9‑97

The energy balance leaf temperature Teb is then

9‑98

Sensor head calibration coefficients

Table 9‑3. Sensor head calibration coefficients, with XML location.
Symbol Description XML Locator (/licor/li6850/...)
af1...af7
ar1...ar4

as1...as4
Main flow sensor
Ref flow sensor
Sample flow sensor
../factory/flowmeter/a...g
../factory/irga_a/flow/a1...a4
../factory/irga_b/flow/a1...a4
Bwc
Boc

Bow1
,
Bow2
Band broadening coefficient for water on CO2
Band broadening coefficient for oxygen on CO2
Band broadening correction for oxygen on water
../factory/bb/ch
../factory/bb/cx
../factory/bb/hx0, hx1
br1...br5
bs1...bs5
Reference CO2 calibration coefficients
Sample CO2 calibration coefficients
../factory/irga_b/co2/a1...a5
../factory/irga_a/co2/a1...a5
dr1...dr3
ds1...ds3
Reference H2O calibration coefficients
Sample H2O calibration coefficients
../factory/irga_b/h2o/a1...a3
../factory/irga_b/h2o/a1...a3
mco...mc3
mwo...mw3
CO2 match coefficients
H2O match coefficients
../cfg/match/co2_adj ... co2 adj 3
../cfg/match/h2o_adj ... h2o adj 3
pca
pcb
Pressure correction or fan speed and flow rate
Pressure correction or fan speed and flow rate
../factory/chamber/pca
../factory/chamber/pcb
σcr1...σcr3
σcs1...σcs3
σwr1...σwr3
σws1...σws3
CO2 reference absorptance span drift with temp
CO2 sample absorptance span drift with temp
H2O reference absorptance span drift with temp
H2O sample absorptance span drift with temp
../factory/irga_b/co2/s1...s3
../factory/irga_a/co2/s1...s3
../factory/irga_b/h2o/s1...s3
../factory/irga_a/h2o/s1...s3
Scro
Scr1
Scso
Scs1
Span1 for reference CO2
Span2 for reference CO2
Span1 for sample CO2
Span2 for sample CO2
../cal/co2bspan1
../cal/co2bspan2
../cal/co2aspan1
../cal/co2aspan2
Sqin
Sqout
In-chamber light sensor cal
External quantum sensor cal
../cfg/ppfdin/mult
../cfg/ppfdout/mult
Swro
Swr1
Swso
Sws1
Span1 for reference H2O
Span2 for reference H2O
Span1 for sample H2O
Span2 for sample H2O
../cal/h2obspan1
../cal/h2obspan2
../cal/h2oaspan1
../cal/h2oaspan2
Vfo
Vfro
Vfso
Zero offset for main flow meter
Zero offset for reference flow meter
Zero offset for sample flow meter
../cal/flowmeterzero
../cal/flowbzero
../cal/flowazero
Vpo Zero parameter for differential pressure sensor ../cal/chamberpressurezero
Xo Oxygen concentration, percent ../factory/cal/oxygen
Xcwr
Xcws
Xwcr
Xwcs
Cross sensitivity, CO2 on H2O, reference cell
Cross sensitivity, CO2 on H2O, sample cell
Cross sensitivity, H2O on CO2, reference cell
Cross sensitivity, H2O on CO2, sample cell
../factory/irga_b/xhc
../factory/irga_b/xch
../factory/irga_a/xhc
../factory/irga_a/xch
Zcr
Zcro
Zcs
Zcso
Zero drift with temperature for reference CO2
Zero offset for reference CO2
Zero drift with temperature for sample CO2
Zero offset for sample CO2
../factory/irga_b/z
../cal/co2bzero
../factory/irga_a/z
../cal/co2azero
Zwr
Zwro
Zws
Zwso
Zero drift with temperature for reference H2O
Zero offset for reference H2O
Zero drift with temperature for sample H2O
Zero offset for sample H2O
../factory/irga_b/z
../cal/h2obzero
../factory/irga_a/z
../cal/h2obzero

Sensor measurements and computations

Table 9‑4. Sensor measurements and computations, and where to find them in the Data Dictionary.
Symbol Description (units) Name (Label) Group
αcr
αcs
αwr
αws
Reference cell CO2 absorptance
Sample cell CO2 absorptance
Reference cell H2O absorptance
Sample cell H2O absorptance
abs_c_b
abs_c_a
abs_h_b
abs_h_a
Raw
Raw
Raw
Raw
co...c4 Boundary layer function coeffs blc_a...blc_e ChambConst
Ca
Cb
Cr
Crd
Cs
Csd
∆Pc
Sample cell CO2, not adjusted for match
Sample cell CO2 (µmol mol1)
Reference cell CO2 (µmol mol1)
Reference cell CO2, dry mole fraction
Sample cell CO2 (µmol mol1)
Sample cell CO2, dry mole fraction
Chamber over pressure (kPa)
CO2_a
CO2_b
CO2_r
CO2_r_d
CO2_s
CO2_s_d
Pchamber (∆Pcham)
Meas
Meas2
Meas
Meas
Meas
Meas
Meas
er
es
Icr
Icro
Ics
Icso
Reference cell vapor pressure (kPa)
Sample cell vapor pressure (kPa)
Reference CO2 raw detector count
Reference CO2 raw detector reference count
Sample CO2, raw detector count
Sample CO2 raw detector reference count
e_r
e_s
Wc_r
Wco_r
Wc_s
Wco_s
Meas2
Meas2
Raw
Raw
Raw
Raw
Iqin
Iqout
In-chamber PPFD sensor raw counts
External quantum sensor raw counts
   
Iwr
Iwro
Iws
Iwso
Reference H2O raw detector count
Reference H2O raw detector reference count
Sample H2O raw detector count
Sample H2O raw detector reference count
Ww_r
Wwo_r
Ww_s
Wwo_s
Raw
Raw
Raw
Raw
Pa Atmospheric pressure (kPa) Press (Pa) Meas
F
Fr
Fs
Flow to chamber (µmol s1)
Flow from reference cell (µmol s1)
Flow from sample cell (µmol s1)
Flow
Flow_r
Flow_s
Meas
Status
Status
G Chamber fan rotation rate (rpm) Fan speed Status
Qin
Qout
In-chamber PPFD
External PPFD
PPFD_in (Q_amb_in)
PPFD_out (Q_amb_out)
Meas
Meas
Ta Leaf chamber air temperature (C) Tchamber (Tair) Meas
Tdr
Tds
Dewpoint temperature reference cell (C)
Dewpoint temperature sample cell (C)
Td_r
Td_s
Meas2
Meas2
Tj1
Tj2
Tk
Leaf T/C 1 junction temperature (C)
Leaf T/C 2 junction temperature (C)
IRGA block temperature (C)
Tleafjunction (Tleaf_j)
Tleafjunction2 (Tleaf2_j)
Tirga_block (Tirga)
Status2
Status2
Status
Tl1
Tl2
Tr
Ts
Leaf temperature 1 (C)
Leaf temperature 2 (C)
Reference cell inlet temperature (C)
Sample cell inlet temperature (C)
Tleaf
Tleaf2
Tb (Tr)
Ta (Ts)
Meas
Meas
Status
Status
Vf
Vfr
Vfs
Signal (V) from main flow sensor
Signal (V) from reference flow sensor
Signal (V) from sample flow sensor
Flow
Flow_b_v (Flow_r_v)
Flow_a_v (Flow_s_v)
Raw
Raw
Raw
Vl1
Vl2
Leaf temperature 1 signal (mV)
Leaf temperature 2 signal (mV)
leaf_t_mv (Tleaf_mv)
leaf2_t_mv (Tleaf2_mv)
Raw
Raw
Vp Differential pressure signal (V or kPa) VPchamber Raw
Wa
Wb
Wr
Ws
Sample cell H2O not corrected for match
Reference cell H2O
Reference cell H2O (mmol mol1)
Sample cell H2O (mmol mol1)
H2O_a
H2O_a
H2O_r
H2O_s
Meas
Meas2
Meas
Meas