Other software settings

This section describes features that have not been described elsewhere.

Flow module options

Under the flow module options, you can set the flow rate to the recommended 15 LPM and review other parameters.

  • Measured flow (SLPM): Standard liters per minute.
  • Measured flow (LPM): Liters per minute; volumetric flow corrected for temperature and pressure).
  • Pressure drop (kPa): An indicator of blower pressure. Values can range from 0 to 4.5, with 1.0 being typical. The value reflects resistance to flow, which can be caused by filters, collapsed tubing, or other obstructions. Record this value as a baseline when you first install the instrument.
  • Motor drive (%): The power supplied to the motor. Typically near 25%.
  • Flow: On or Off. This option controls whether the flow module is on or off.
  • Address: The address setting must match the address displayed inside the flow module case. If the addresses do not match, change one or the other so they are the same. Addresses are 32 to 255. Normally, no action is required.
  • Set the flow rate in SLPM: We recommend this setting with a flow rate of 15.0 SLPM. The slider displays values between 10 and 18 SLPM. During typical operation at 15.0 SLPM, the blower motor drive runs at about 25%. Measured flow is still given in SLPM or LPM.
  • Set the flow rate to a motor speed: When configured for motor speed, the slider displays 0 to 100%. Motor speed % is not recommended for eddy covariance applications.

Flow module interior control panel

The flow module has a control panel and LED indicators. We recommend controlling the flow module using the software interface, but the hardware controls are available never-the-less.

  • Measured Flow (SLPM): The current flow rate in mass flow units of Standard Liters Per Minute.
  • Set Flow (SLPM%): When selected, allows you to set the desired flow rate by turning the Adjust know. Flow is displayed as a constant blower drive mode given as a percentage of the total flow capacity. If you turn the Adjust knob past the maximum of 18.0 SLPM, the display will change to show a C for constant, and the displayed setting is constant blower motor drive in % air flow.
  • Pressure Drop (kPa): An indicator of pressure loss due to tubing, filters, the sensor head, and any obstructions. The range is from 0-4.5 kPa, with a normal reading of about 1.0 kPa. If the air filter becomes clogged, this number will increase, which is a useful diagnostic indicator flow performance.
  • Motor Drive (%): Displays the power to the blower motor as a percentage. During normal operation, the blower motor will operate at about 25% drive. This value will increase if flow is constricted.
  • Address: Displays the accessory bus address (between 32-255). This address must match the address in 7200 > Flow Module tab in the software.
  • Power Save: After about 7 minutes of control panel inactivity, the control panel will enter a power save mode, leaving only the Power Save LED illuminated. Press any other button to exit power save mode.
  • Power: On when power is applied to the instrument.
  • In Memory: On when the displayed value matches the value stored in memory. The two parameters that can be stored are the Flow setting and the Address. Whenever the Set Flow LED is selected, the flow setting is recalled from memory and displayed, and the In Memory LED will be lit. If the Adjust knob is turned, the In Memory LED will remain off until the Store In Memory button is pressed or the Adjust knob is turned until the displayed setting once again matches the stored setting. This same behavior applies the Address LED.
  • Store In Memory button: The flow module has built in memory for retaining the Set Flow and Address settings. These values are retained in memory after the 7200-102 is powered off. If you are changing a setting with the knob, when the desired setting is displayed, press Store In Memory to save the setting.
  • Flow ON: Indicates whether the blower motor is on or off.
  • Flow ON/OFF button: Turns the blower motor on or off.
  • Menu button: Press once to change the displayed parameter. The displayed parameter is indicated by the illuminated LED.

Settings

Under Settings, you can set the instrument time, network configuration, send commands, change the chopper housing temperature set point, and integrate CO2 measurements.

Time

This is where you set the instrument time and date. The instrument clock uses the Precision Time Protocol (PTP) time keeping system. PTP is a high precision time synchronization protocol for networked devices. Devices controlled with PTP can maintain accuracy in the sub-microsecond range with a sufficiently accurate master clock. PTP is defined in the IEEE 1588-2002 and 1588-2008 standards, entitled “Standards for Precision Clock Synchronization Protocol for Networked Measurement and Control Systems.” A detailed summary of IEEE-1588 is available at www.ieee1588.com. Full documentation is available for purchase from the Institute of Electrical and Electronics Engineers (IEEE) at www.ieee.org.

The basic principle behind PTP is that the best time keeping can be accomplished with multiple networked devices by synchronizing all device clocks to the most precise clock on the network. Each clock on the network has a rating that indicates its relative accuracy. The IEEE 1588 protocol specifies the use of a Best Master Clock algorithm to determine which clock on the network is the most accurate. On a network, the most accurate clock becomes the master clock and all other clocks sync to the master clock.

The software implementation of PTP provides accuracy in the 10 microsecond range. When used with the SmartFlux System, the GPS clocks will become the master clock for the system.

About time keeping

The LI-7500A/RS/DS and LI-7200/RS are network-based instruments and it is possible for multiple users to log data from a single instrument over multiple TCP/IP connections. Consequently, the analyzer uses Coordinated Universal Time (UTC) for its onboard timekeeping tasks. The default time stamp is therefore UTC based, but local time can be set, if desired.

Generally, we recommend that the system synchronizes its time to the GPS clocks. The instrument uses a time zone database that includes local time zones that are kept as constant offset from UTC. These time zones are listed as ‘Etc/GMT’ + offset. However, these time zone names beginning with ‘Etc/GMT’ have their sign reversed from what is commonly used. Thus, zones west of GMT have a positive sign, and those east have a negative sign. For example, US/Central Standard Time is 6 hours behind GMT, and in the database this time zone is listed as ‘Etc/GMT-6’.

Unix time is the number of seconds elapsed since the Unix epoch of 00:00 Coordinated Universal Time (UTC) January 1, 1970 (or 1970-01-01T00:00:0Z ISO 8601). For example, the time stamp 1262884605 translates to 01/07/2010 at 05:16:45 UTC. The date and time are converted to a conventional display format (YYYY-MM-DD; HH:MM:SS) and adjusted based on the time zone setting that you select.

The time stamp in each file header shows the instrument time and time zone.

Setting the clock

  1. Connect to the gas analyzer.
  2. Open the Settings window.
  3. Set the Clock Sync (PTP).
  4. Choose your time zone.
  5. Click Apply or OK.

The PTP clock settings available are:

  • Off: Turns PTP time keeping system off. Instrument time will be determined by the Date and Time set by the user, even if there is a better clock on the network.
  • Automatic: The gas analyzer searches the network and syncs to the most accurate clock using the Best Master Clock algorithm (could be the gas analyzer). This setting should be used in most circumstances.
  • Slave Only: The gas analyzer always syncs to another clock. It will search the network and synchronize to the best clock.
  • Preferred: The gas analyzer uses its own internal clock unless it finds a better clock on the network.

Network

The instrument name is given in the Connect window and Settings > Network window. You can change this name in Settings > Network. The name can include upper and lower case letters (A to Z; a to z), numbers (0 to 9), dash (-), and period (.). Other characters are invalid and will cause communication errors with the system.

The IP address (Internet Protocol) is a numerical identifier that is assigned to devices participating in a computer network. In many cases this address is assigned automatically (Dynamic IP address). In other cases, your network administrator may assign a permanent address (Static IP address) that can be entered manually.

With Obtain an IP address automatically selected, the IP address, Subnet mask, and Default Gateway fields are filled in automatically; if you choose to enter the IP address manually, the address fields are editable.

The Subnet mask is a set of 4 octets used to separate an IP address into two parts; the network address and the host address. The Gateway is a node that routes traffic to another network.

Manual addressing

You can connect to an instrument by manually enter the IPv4 Address or Hostname of the instrument. The IP address can be set automatically by a network or assigned manually in Settings > Network > Use the following IP address. If IPv6 (Internet Protocol version 6) addressing is available on the network and enabled on the computer, the instrument’s IPv6 address is displayed as well.

Both the LI-7500A/RS/DS and LI-7200/RS use the Port number 7200 for Ethernet communications. Therefore, when connecting to the instrument on a network you will enter the IP address of the instrument and port 7200 in the connection window of the software to initiate communication. You can select the instrument from the list of instruments on the same network as your computer or connect your computer directly to the instrument Ethernet port.

Port forwarding

In some network setups it may be necessary to forward communication traffic on a port from a public IP address to a private IP address to gain access to an instrument. For example, assume you have an analyzer installed in the field and the instrument is connected to a wireless gateway such as a cellular modem. The instrument will acquire a private IP address from the cellular modem, but this address is only visible to nodes on the private network. On the other hand, the cellular modem is assigned a public IP address that can be accessed from any node on the Internet. For this discussion let’s assume the public address is 166.66.77.88. In order to connect to the instrument in the private network, a port forwarding rule must be created in the cellular modem to “forward” all communications on port 7200 coming into the modem to the private IP address of the instrument.

Instrument Example Public IP Public Port Private IP Private Port
1 166.66.77.88 7200 192.168.13.201 7200

The public port number should be changed when there is more than one LI-7500A/RS/DS or LI-7200/RS instrument in the private network. In this case, to connect to each instrument, two port forwarding rules must be set up, similar to that shown below.

Instrument Example Public IP Public Port Private IP Private Port
1 166.66.77.88 7201 192.168.13.201 7200
2 166.66.77.88 7202 192.168.13.202 7200

You can run multiple LI-7500A/RS/DS or LI-7200/RS software sessions at the same time to communicate with different analyzers. Simply double-click on the software icon to open another session, and connect to a different instrument.

Manual

A command line field is present that allows you to send a command to the instrument. This can be useful for diagnosing problems, as a LI-COR technician can gauge the instrument’s response to given commands, and determine if the instrument is functioning properly. Contact LI-COR technical support for details on the grammar.

Advanced—Chopper housing temperature

The chopper motor housing temperature can be set to a lower operating temperature (5 °C) in winter to reduce power consumption and minimize heating by the electronics. We recommend changing the setting only when the average ambient temperature drops below 5 °C. You can change the setting back to 30 °C when the average ambient temperature is above 5 °C. Note, however, that the instrument will still function properly when the chopper motor housing temperature is set to 30 °C, even when temperatures are below 5 °C.

Do not set the chopper housing temperature to 5 °C when ambient temperatures are above 5 °C, however, as the instrument will not function properly.

Important: When changing between winter and summer settings, you will need to perform a zero and dual span calibration.

Integration

Allows for the setup of an integration function using a selected CO2 source (mmol/m3, µmol/mol, or dry µmol/mol). The integrated value selected can be viewed as a data source in the main software window (choose Integral in the Data Items list). To integrate a CO2 source:

  1. Choose a source to be integrated from the Integration Channel list.
  2. Choose the method to start the integration; immediately, or using a threshold value.
  3. Choose the method to stop the integration; manually, with a threshold value, or after a user-entered elapsed time has expired.
  4. Enter the threshold value for the start or stop (or both) of the integration, if either was chosen in Steps 2 or 3 above.
  5. Click on the Apply button.
  6. If Start Immediately was chosen for the Start Time, the integration function is started and/or stopped manually by clicking the òdt Integration On or òdt Integration Off buttons (available on the dashboard, as well).
  7. Example: Start integrating CO2 (µmol/mol) after it reaches a value of 500 µmol mol-1, and continue integrating for 5 minutes.
  8. Choose CO2 (µmol/m3) from the Integration Channel list.
  9. Click Start when channel reaches button and enter 500 in the text entry field.
  10. Click the Stop after an elapsed time button and enter 300 (seconds) in the text entry field.
  11. Click Apply.
  12. Click the òdt Integration On button. Click OK to dismiss the dialog.

The integration result (area under the curve) can be viewed in the dashboard.

The LI-7200RS menu provides access to Inputs, Outputs, Calibration, PC Logging, Flow Module, and Intake Tube settings.

Inputs

There are three things to configure under Inputs:

  • Auxiliary Inputs: Can be used for sonic anemometer analog data.
  • Temperature Source: Specifies whether the temperature source is the instrument, a temperature sensor on Auxiliary Input 1, or a User Entered value.
  • Pressure Source: Specifies whether the temperature source is the instrument, a pressure sensor on Auxiliary Input 2, or a User Entered value.

Temperature and pressure values are required to convert CO2 and H2O density (mmol/m3) to mole fraction (µmol/mol or mmol/mol). In addition, the analyzer requires a pressure value to compute CO2 or H2O mole density, and a temperature value to perform the band broadening correction for H2O on CO2.

In eddy covariance applications we recommend that you use temperature and pressure measurements provided by the instrument.

Note: Pressure is measured at both the LI-7200RS sensor head, and in the LI-7550 Analyzer Interface Unit. The LI-7200RS sensor head pressure is displayed as the differential between the head and the LI-7550. The Total Pressure is the box pressure plus the head pressure.

If you use your own temperature and/or pressure sensors, one or two of these channels will be unavailable for anemometric data. If you use the on-board temperature and pressure sensors, all four auxiliary input channels are available for use with other sensors.

Choose Other for any sensor input other than a sonic anemometer. The Units field describes the data after the coefficients m and b are applied. The Label will appear in the header of the data file.

Each auxiliary input channel has two fields for the multiplier (m) and offset (b). Linear voltage inputs (0 to 5 V) use the slope-offset equation:

10‑1y = m*x + b

where y is the sensor output, x is the voltage output of the sensor, b is the Y-axis intercept (offset), and m is the calibration multiplier, which is the slope of the line representing the sensor’s response.

Outputs

Five options are available under the outputs tab: Setup, RS-232, Ethernet, DACs, and SDM..

Setup

The Setup tab presents Delay Time and Bandwidth settings that affect all outputs (RS‑232, Ethernet, DACs, and SDM).

Output bandwidth

Bandwidth (5, 10 or 20 Hz) determines the signal averaging done by the digital filter. To avoid aliasing (only a concern for co-spectra, not for fluxes), one should sample the gas analyzer at a frequency greater than or equal to 2 times the bandwidth. Thus, if you are sampling at 10 Hz, the bandwidth is 5 Hz.

Bandwidth is the frequency at which the indicated amplitude is 0.707 of the real amplitude (Figure 10‑3).

Figure 10‑3. Bandwidth = 1/oscillation period.

Bandwidth is a useful indicator for characterizing real-world behavior in which there are fluctuating gas concentrations. Given a sinusoidal oscillation of concentration, the instrument's ability to measure the full oscillation amplitude diminishes as the oscillation frequency increases.

Note: The bandwidth selection has no impact on the system delay. The filters were designed so they have exactly the same delay whether a 5, 10, or 20 Hz signal bandwidth is selected.

RS-232 output

The RS-232 tab is used to set the instrument's RS-232 port configuration for unattended data collection. After configuration, click Apply. The instrument will begin to send data out the RS-232 port according to these parameters after you disconnect from the instrument (or immediately if you are connected via Ethernet).

Use the Baud Rate menu to select from 9600, 19200, 38400, 57600, or 115200 baud.

Use the Update Rate menu to select from .1, .2, .5, 0, 1, 2, 5, 10, or 20 Hz. Note that at lower baud rates, you may not be able to output all data items at high frequencies. For example, at 9600 baud, the maximum update rate is 2 Hz, when outputting all 34 data items.

Note on Data Output Rates

When sending data via a serial connection (e.g., RS-232), note that the baud rate selected may limit the number of samples that can be output. A single line of data (all log values selected) consists of approximately 350 bytes (2800 bits), so the maximum output rate at the available baud rates becomes:

Baud Rate Maximum Output Rate (Samples per Second)
9600 2
19200 5
38400 10
57600 15
115200 20

Under Options, you can choose whether or not to output labels with each data record, and whether to output diagnostic text records. An example of a data record sent with and without labels is shown below.

Data format with labels:

(Data (Ndx 87665) (DiagVal 757) (Date 2009-09-10) (Time 14:06:44:140) CO2Raw 0.0332911) (H2ORaw 0.19299) (CO2D 5.20672) (H2OD 755.566) (Temp 15.517) (Pres 99.4361)…

Data format without labels:

87665 757 2009-09-10 14:06:44:140 0.0332911 0.19299 5.20672…

The End of Line Delimiters determine the character(s) that terminate the data records. Depending on your recording device, a line feed, carriage return, or both may be required to properly parse the data records.

Under Data Output, select the data records that you want to output; click Select All to choose all, or Select None to disable all checked values.

Ethernet output

The Ethernet tab is used to set the instrument's Ethernet output. After configuration, click Apply and the analyzer will begin to send data out the Ethernet port according to these parameters.

Use the Update Rate menu to select from 0.1, 0.2, 0.5, 0, 1, 2, 5, 10, or 20 Hz.

Under Options, you can choose whether or not to output labels with each data record, whether to output diagnostic text records, and whether to use verbose mode. An example of a data record sent with and without labels is shown below.

Data format with labels:

(Data (Ndx 87665) (DiagVal 757) (Date 2009-09-10) (Time 14:06:44:140) CO2Raw 0.0332911) (H2ORaw 0.19299) (CO2D 5.20672) (H2OD 755.566) (Temp 15.517) (Pres 99.4361)…

Data format without labels

87665 757 2009-09-10 14:06:44:140 0.0332911 0.19299 5.20672…

Under Data Output, select the data records that you want to output; click Select All to choose all, or Select None to disable all checked values.

DAC configuration

The LI-7550 has the capability to output up to 6 variables on DAC channels 1-6. The DACs page allows you to configure the DAC output channels by specifying the source channel (e.g., CO2 mmol/m3) that drives the analog signal, the source channel value that corresponds to zero volts, and the source channel value that corresponds to full scale voltage (5V).

For example, to configure DAC #1 to output a voltage signal proportional to CO2 mmol/m3, 20 mmol/m3 full scale, select CO2 mmol/m3 under DAC1 'Source', and set 0V = 0, and 5V = 20.

0V → 0(Xo, zero volts corresponds to 0 mmol/m3)

5V → 20(XF, full scale corresponds to 20 mmol/m3)

When a voltage range R is selected, the DAC output voltage V resulting from a CO2 molar value X is given by

10‑2

where R = 5V.

The DACs are linear, so in the example above, a measured voltage signal of 3 volts would correspond to a CO2 mmol/m3 value of 12.

For test purposes, you can also choose Set Point for the Source, and enter a Set Point voltage value; the DAC channel will then output that voltage continuously.

SDM output

SDM addressing allows multiple SDM-compatible peripherals to be connected to a single Campbell Scientific datalogger. Choose an address under the SDM tab between 0 and 14.

The pin assignments of the SDM Interface Cable are given in Pin assignments. Generally, the blue wire (SDM_DATA) will connect to C1 on the datalogger, the white wire (SDM_CLK) to C2, the brown wire (SDM_EN) to C3 and the black wire (ground) to ground.

SDM communications are enabled in Campbell Scientific®, Inc. dataloggers with the LI7200 instruction. This instruction is only available in dataloggers programmed in CRBasic (e.g., CR5000). When using this instruction parameter three (SDMAddress) should be set to match the SDM address of the LI-7200RS. Parameter four (LI7200Cmd) is used to define what variables the datalogger collects from the instrument. The LI7200 instruction supports a group trigger mode, and so it can take on values from 0 to 7 (see Table 10‑1 for variables are included in each option).

Below is a programming example of the LI7200 instruction used to collect data from an LI-7200RS at SDM address 0:

LI7200 (LI-7200(),1,0,6)

Table 10‑1. Parameter 4 value definitions.
Mode Items Sent
0 CO2 dry (µmol/mol)
H2O dry (mmol/mol)
1 CO2 (mmol/m3)
H2O (mmol/m3)
Ptotal (kPa)
Tin (°C)
Tout (°C)
2 CO2 dry (µmol/mol)
H2O dry (mmol/mol)
CO2 (mmol/m3)
H2O (mmol/m3)
Signal Strength (average)
Ptotal (kPa)
Tin (°C)
Tout (°C)
Cell Temperature (°C)Aux channel #1
3 CO2 dry (µmol/mol)
H2O dry (mmol/mol)
CO2 (mmol/m3)
H2O (mmol/m3)
Signal Strength (average)
Head Pressure (kPa)
Tin (°C)
Tout (°C)
Aux channel #1
Aux channel #2
Aux channel #3
Aux channel #4
4 CO2 (mmol/m3)
H2O (mmol/m3)
CO2 absorptance
H2O absorptance
Signal Strength (average)
Ptotal (kPa)
Tin (°C)
Tout (°C)
Block Temperature (°C)
Diagnostic value (see below)
Aux channel #1 (user units)
5 CO2 dry (µmol/mol)
H2O dry (mmol/mol)
Signal Strength (average)
Cell Temperature (°C)
Ptotal (kPa)
6 Diagnostic Value (see below)
7 Trigger Mode
CO2 dry (µmol/mol)
H2O dry (mmol/mol)
Signal Strength (average)
Ptotal (kPa)
Cell Temperature (°C)
NOTES: LI-7200RS Mode 7 requires the Campbell datalogger to broadcast a group trigger to cause data to be registered. The registered data set is held and not updated until the next group trigger.

Delay time

The output signal from the optical bench is sampled by a high-speed analog-to-digital converter and input into a digital signal processor (DSP). This signal is processed digitally and gas densities are computed from it. There is a fixed delay in this process, and an additional user-programmable delay that can be used to make the instrument output occur on even sampling intervals.

The instrument has a fixed throughput delay of 172 milliseconds for the DAC outputs, and 130 ms for the SDM, RS-232, and Ethernet outputs. This delay can be increased in increments of 1/150 seconds (6.667 ms), to minimize offsets between the gas analyzer and other sensors.

For example, suppose you are sampling (via SDM output) the LI-7500A/RS or LI-7200/RS with a Campbell Scientific CR5000 datalogger at 10 Hz (0.1 s). Setting the delay count of the gas analyzer to 25 yields a total delay of 0.297 seconds, which means the data will have a delay of 3 execution intervals (0.297 s/0.1 s), which the logger can allow for when synchronizing the data to the sonic anemometer or other analog measurements made by the datalogger. Thus, the “unaccounted for” lag is 0.003s. Without this extra delay, the lag time would be 0.130 seconds, which is 1 execution interval (0.1 seconds) plus 0.03 seconds unaccounted for.

Similarly, if you are sampling the gas analyzer with the DAC outputs, setting the delay count to 19 yields a total delay of 0.299 seconds. The lag will be 0.001s.

Total system delay examples

Total System Delay (ms) = Delay Time + (Delay Step × Delay Step Increment)

Output Delay Time (ms) Delay Step (ms) Delay Step Increment Total Delay (ms)
DAC 172 6.667 19 299
SDM/RS-232/Ethernet 130 6.667 25 297

Calibration window

The Calibration window is where you set the zero and span of the instrument. There are entry fields to set the target values for the span gases used to set the span of the instrument; CO2 span gas target values are in ppm, and H2O span gas target values are entered in °C dewpoint. The Zero and Span tabs also provide information about the stability of the gas flowing through the optical path:

  • A green flag indicates that it is OK to perform the calibration
  • A red ‘X’ indicates that it is not OK to perform the calibration; wait until the red ‘X’ changes to a green flag before performing the calibration.

Zero, Span, and Span 2 tabs

Step-by-step instructions for calibrating the gas analyzer can be found in User calibration.

Signal strength tab

The Signal Strength tab has a button labeled Reset Signal Strength that you can click if you’ve decided your instrument optics are as clean as you can reasonably make them, and you want to reset the signal strength to 100.

See CO2 signal strength for more information.

Coefficients tab

The Coefficients tab displays factory-determined calibration coefficients, a factor for correcting CO2 measurements for band broadening due to the presence of water vapor (A), and a zero drift correction factor (Z). The coefficient shown as XS (cross sensitivity) compensates for slight cross sensitivity between CO2 and H2O signals absorbed by the detector.

The calibration coefficients, XS, and Z values are unique to each sensor head, and may be found on the calibration sheet shipped from the factory. The Band Broadening coefficient is 1.15 for all sensor heads.

The calibration coefficients are stored in the LI-7550 Analyzer Interface Unit, not the sensor head. To change these values, edit them and then click Apply to load these values to the LI-7550.

Caution: To avoid undesirable changes to the instrument performance, do not edit values in this window. Click Cancel to avoid any undesirable changes.

The Head Serial Number displays the serial number of the sensor head that is associated with the coefficients. When exchanging sensor heads, it is necessary to change calibration coefficients, and to redo the zero and span.

Note: The easiest way to get the calibration coefficients, zero, and span updated for a sensor head is to save the configuration (it is a *.7x file) and then upload the file to a different LI-7550 using Open Configuration.

Manual tab

The Manual tab displays current values for CO2 and H2O zero and span; after performing a zero and span calibration, these values should be near 1 (zero is typically between 0.8 and 1.2, and span is typically between 0.8 and 1.2; Zwo and Zco, see equations 11‑13 and 11‑17). Span 2 values will be near zero.

We recommend that you track the zero values over time as you re-zero the instrument. As the internal chemicals lose their effectiveness, this value will increase. The CO2 zero drift is also somewhat temperature dependent.

In most cases you will never need to manually edit these values. If, for some reason, the instrument calibration becomes unstable (e.g., you accidentally zeroed or spanned the instrument with the wrong gas), you can manually enter a value of 1 for each of these parameters (and zero for Span 2 values), and click Apply. This will return the instrument to a more normal state, after which you can perform zero and span calibrations again. Alternatively, if you have performed at least one previous (successful) calibration, you can restore those values using the History tab.

History tab

The History tab displays a list of previous calibration backup files generated during a zero and/or span calibration of the instrument (when used with the same computer used to connect to the instrument). You can click on any previous calibration in the list to view the details; you can also restore the values from a previous calibration by choosing a file in the list and clicking the Restore button. Click Delete to permanently remove calibration files from the list stored on the computer.

PC logging

The PC Logging window is used to configure the data output parameters used while the LI-7500A/RS/DS or LI-7200/RS Windows software is active. Data are logged to a file on your computer. You can specify the destination file, update rate, file duration, comments, and values to be logged.

Data can be logged at up to 20 Hz. These files can be split into smaller files, at 15, 30, 60 minutes, or 1.5, 2, 4, or 24 hour intervals. The files are split based on fractions of the hour as measured by the system clock. Thus, if you choose to split the files at 15 minute intervals and start logging at 10:22, the file will be split at 10:30, 10:45, 11:00, etc.

Note: File compression and metadata information are not available when logging to a PC; these options are available only when logging to the USB drive.

Under Log Values, select the data records that you want to output; click Select All to choose all variables.

Click Start to begin logging data and Stop to quit logging.

Note on data output rates

When sending data via a serial connection (e.g. RS-232), note that the baud rate selected will limit the number of samples that can be output. A single line of data (all Log Values selected) consists of approximately 350 bytes (2800 bits), so the maximum output rate at the available baud rates becomes:

Baud Rate Maximum Output Rate (Samples per Second)
9600 2
19200 5
38400 10
57600 15
115200 20

Logging 20 samples per second (20 Hz) would result in approximately 384 MB/day data accumulation, or about 1 GB/2.5 days.

Diagnostics

The Diagnostics page allows you to view the current operational state of the instrument, including current software versions, diagnostic flags, optical bench properties, and technician test point values.

Status tab

Diagnostic indicators for the Sync Flag, PLL, Detector Temperature, and Chopper Temperature will read either OK or Service.

Instrument Software: Reports the Embedded version, DSP version, and FPGA version. For diagnostic purposes.

Head: Displays the model of the sensor head currently connected to the LI-7550 Analyzer Interface Unit. The software may display LI‑7200 even if an LI-7200/RS is connected and LI‑7500A even though an LI-7500A/RS is connected.

Avg. Signal Strength: Average Signal Strength is an indicator of the cleanliness of the optics. A value near 100 indicates clean optics; a value near 0 indicates highly contaminated optics. For the best performance, clean the optics if the signal strength is below the upper 90s. Signal strength is not an indicator of accuracy because contamination on the optics can affect measurements in unpredictable ways, depending upon the optical properties of the contaminants.

Sync Flag: Not used. Will always read OK.

PLL: Phase Lock Loop offset, indicates the status of the optical filter wheel for the IRGA. Common causes of a service indication are not having a sensor head connected to the LI-7550, the head cable is not been tightened properly, or the cable has failed.

Detector Temperature: If not OK, indicates the detector cooler is not maintaining the proper temperature. This will happen at temperatures above 50 °C. This does not always indicate a serious problem; the cooler may simply have not yet reached the target temperature during instrument start up, or it may be out of range due to external environmental conditions. Readings may still be OK. Otherwise, this can occur if the sensor head is not be attached to the LI-7550 or if the cable is not tight, causing a bad connection.

Chopper Temperature: If not OK, indicates the chopper temperature controller is out of range, hot or cold. As with the Detector Temperature indicator above, this may or may not indicate a serious problem. The chopper should be able to temperature control when ambient is between +50 and -25 °C. Check that the sensor head cables to the LI-7550 are tight, and make sure the chopper temperature is not set to Winter when measuring in high ambient temperatures.

Aux Inputs: If not OK, indicates that the two reference voltages on the LI-7550 analog circuit board cannot be checked properly. The LI-7550 needs service.

Head Pressure: If not OK, the pressure reading at the sensor head is outside of its normal range. A service indication can also be caused by a loose or missing cable between the LI-7550 and the sensor head.

Inlet T- If not OK, the temperature reading at the sensor head inlet is outside of its normal range. This can be caused by a loose or missing cable between the LI-7550 and the LI-7200/RS sensor head, or the inlet thermocouple has been broken.

Outlet T - If not OK, the temperature reading at the sensor head outlet is outside of its normal range. This can be caused by a loose or missing cable between the LI-7550 and the LI-7200/RS sensor head, or the outlet thermocouple has been broken.

Test point values

The Test Points tab displays voltages and raw counts of a variety of diagnostic test points. Though primarily for LI-COR technician reference, values outside of the normal range may give you an indication of where problems may be originating.

Test Point Description Normal range
IR A/D (V) Voltage from IRGA being sent to A/D converter 0.2 – 4.8 V
Detector Temp (V) Detector temperature voltage 2.9 V. Readings near +5 V or 0 V are normal for a few seconds after power on.
Aux Ref + 16-bit conversion of positive reference voltage used in Auxiliary Input circuit ~64830
Aux Ref - 16-bit conversion of negative reference voltage used in Auxiliary Input circuit ~600
ADC CH1 (V) Analog input channel 1 -5 V to +5 V. A reading near 0 is normal for an open input.
ADC CH2 (V) Analog input channel 2 -5 V to +5 V. A reading near 0 is normal for an open input.
ADC CH3 (V) Analog input channel 3 -5 V to +5 V. A reading near 0 is normal for an open input.
ADC CH4 (V) Analog input channel 4 -5 V to +5 V. A reading near 0 is normal for an open input.
Box Pressure (V) Absolute pressure Any value between 0 and 5
Head Pressure (V) Pressure at sensor head 2.5 is normal
In-Temp (V) Temperature at sensor head inlet +2.5 is normal
Out-Temp (V) Temperature at sensor head outlet +2.5 is normal
Box Temp (V) LI-7550 temperature 3.5 at room temperature
Chopper Temp (V) Displays voltage of chopper housing temperature control circuit. 2 is normal. Readings near 0 V or +5 V indicate that the temperature control circuit is unable to control the chopper temperature.
Short (V) Reference short in auxiliary input circuit 32768
Cooler Voltage (V) Displays voltage of detector cooler. If this value gets below 0.5V, it may be too cold; the detector may not be temperature controlling. This will typically occur at -30 °C or below. 0 V to +5 V

Waveform

The Waveform tab displays the current state of the analyzer’s chopping shutter disk. Though primarily for LI-COR technician reference, if problems are encountered, it may be useful to log the waveform data for troubleshooting purposes. Waveform data can only be logged when the Diagnostics window is open.

SmartFlux

SmartFlux System diagnostics are supported by the combined release of win-GHG interface software v8.8.36 and SmartFlux 2/3 System embedded firmware v2.2.50. This feature serves to summarize the system performance by providing status indicators for components and configurations, allowing you to evaluate the system at a glance. Diagnostics are under Diagnostics > SmartFlux tab.

Advanced

The Diagnostics > Advanced tab provides several maintenance and recovery options. These options should be used only after consultation with LI-COR technical support or careful consideration, as all of theses actions except for Reboot SmartFlux will permanently remove data from the device. Once data are removed from the USB device, they cannot be recovered. These actions are performed only on the SmartFlux 2/3 device. Each of these buttons requires confirmation after being clicked.

Important: After clicking any of these buttons, the desktop PC application will disconnect from the instrument. This is confirmation that the command was sent successfully. Do not disconnect manually after clicking; doing so may prevent the command from being sent.

  • Reboot SmartFlux will simply restart the device. Use this if the SmartFlux System is not responding to configuration changes or is not communicating properly. This may cause the Win-GHG application to disconnect (for the SmartFlux 3 system).
  • Erase SmartFlux Onboard Data will clear the internal database, raw data files, results files, and files set for upload from the memory of the SmartFlux System. Use this when the SmartFlux internal database is too large or the internal memory has filled up.
  • Format SmartFlux USB Drive will clear data from the USB drive in the SmartFlux System. Use this when having problems with USB data logging or if manually deleting old files is taking too long.
  • Factory Reset SmartFlux This performs all the items listed under Erase SmartFlux Onboard Data and also removes user configuration such as sonic, and networking. Use this in very rare case where the SmartFlux is malfunctioning and is not able to be configured. The application requires you to wait 15 seconds to contemplate the choice before allowing you to proceed.

Important: If you are moving the SmartFlux 2 from one IRGA to another, you must either Erase SmartFlux Onboard Data or Factory Reset SmartFlux. Use the first option is if you are simply moving the SmartFlux 2 from one IRGA to another that is using the same model of sonic anemometer. Use the second option if you are moving the SmartFlux 2 from one IRGA to another and installing a different model of sonic anemoemter.

Charting

The Chart window allows real time graphics display of one or two variables plotted against time. The Y-Axis displays the value chosen in the respective menu against time on the X-Axis.

Time on the X-Axis can be displayed over user-defined values of seconds or minutes; the X-Axis Max value defines the unit of time displayed in the window before the window starts scrolling the time value off the right edge.

Choose the value for the Y-Axis Left and/or Y-Axis Right in the menu(s), and enter values for the Y-Axis maximum and minimum; choose Auto Scale if you want the chart to scale the Y-Axis automatically to keep data from scrolling off the top or bottom edges. You can also select the color of the line for both Y-Axes values.

When you are finished defining the strip chart(s) parameters, click OK. A new chart window will appear. Press Pause All to temporarily stop plotting (press Pause All again to resume); press Save to save a “snapshot” of the current plot to a bitmap file; press Print to print the currently displayed chart window, or press Options to open the Chart Setup window again to make changes to the strip chart parameters.

Note: You can rescale both Y-axes of an active strip chart by right clicking on the chart, holding, and dragging up or down. You can also open as many chart windows as you want.

Configuration files

The instrument uses a configuration file to store parameters of the configuration, including the setup information, instrument settings, calibration coefficients, and zero and span values, and logging configuration. Unique configuration files can be set up, saved, and then re-opened to easily change your setup information.

When the software program is started for the first time, a default set of parameters is loaded, which can be modified and saved as a new configuration file with a different name. The analyzer stores its configuration so that it will power on configured just as it was when it was powered off.

To save the current configuration, click Config File > Save Configuration. Select the items to include in the file and click Continue. Select a directory and save the file. It will have a .l7x extension.

To open an existing configuration file, click the Config Files button and select Open Configuration. Select the file and click Open. You are prompted to send the configuration to the instrument; changes are implemented in the instrument when the configuration file is loaded.