Connecting to and configuring the system

In this section, we describe how to configure the system using Blueprint Utility Software. You can create a configuration file without connecting to a Data Acquisition Module (DAqM) or Data Retention Module (DRM) but you'll have to connect in order to send the configuration file. If you purchased a sensor package from LI-COR, this section is for you.

Connecting to the biomet system

The Blueprint Utility is software to configure the data acquisition system. Go to licor.com/ec-software for the latest version and install it on your computer.

Here describe how to connect with DAqMs and a DRM, enter sensor calibration information for sensors and send a configuration to the system. Additional information is provided inside the app (click Help) and in Blueprint Utility software reference.

Connecting Ethernet cables

Connect the DRM and all DAqMs together before sending a configuration. You will communicate with the DRM and DAqMs through the SmartFlux 2 or 3 System. The LI-7500DS or LI-7550 (for the LI-7500A/RS or LI-7200/RS) is required to complete the system, but is not required to configure the biomet system. Connect the computer to a vacant port on the network switch.

Power on the system

Move the breaker to the ON position to power on the DRM, DAqM, SmartFlux System, and Network Switch. Allow several minutes for the system to start up.

Launch the Blueprint Utility and load a program

The first time it is opened, the Blueprint Utility presents a blank program. Click Biomet Library and select the basic sensor package.

Connecting to a DAqM and DRM

Click Connect. Identify your SmartFlux System from the list (you'll see the serial number followed by .local, like this: smartx-nnnnn.local, where nnnnn is the serial number). You can also enter the IPv4 or IPv6 address (Windows only) of the SmartFlux System. Click Configure Port and select the SmartFlux port number that the DRM or DAqM is connected to. Click OK.

Configuring Stevens HydraProbes

When connecting for the first time, each HydraProbe must be identified one at a time in the software. If you have already installed the HydraProbe leads, disconnect the power for two of them so that only one is powered on (the one you will designate as Sensor 1).

From Wire Color To
SDI-12 Signal blue DIN Terminal 8
SDI-12 Ground black DIN Terminal 9
SDI-12 12 VDC red DIN Terminal 10

After connecting the first HydraProbe, click SDI-12 Console. Click Query All to make sure that the DAqM sees the sensor and that its address is 0.

Enter the command 0A1 and click Send This command sets sensor 0 address (A) to 1; the default address is 0). The DAqM will respond with 1. Repeat this for the other HydraProbes, connecting the power wire for one sensor at a time and assigning each a unique address (0A2 for sensor 2, and 0A3 for sensor 3).

After each sensor has a unique address, you can set the soil type. For sensor 1, enter 1XW_SOIL_<G/O/R>! (where 1 is the sensor address and <G/O/R> represents the soil type (G for general, O for organic, and R for rockwool). Click Send to send the command. Repeat this for all SDI-12 sensors.

See Overview of SDI-12 commands for more information and details on using HydraProbes with older firmware.

Entering sensor calibration information (Basic Package)

Each program includes some sensors that have specific calibration information. In Blueprint Utility, click the Configuration tab. The required information is described in a note in the program. In this collection of sensors, the soil heat flux plates, NR Lite 2 net radiometer, LI-200R, and LI-190R require calibration information.

Table 4‑1. Constants and the corresponding calibration values that must be entered to use the basic biomet package.
Constant Value to enter
SHF_SENS_1 Soil Heat Flux Plate #1 Sensitivity (S)
SHF_SENS_2 Soil Heat Flux Plate #2 Sensitivity (S)
SHF_SENS_3 Soil Heat Flux Plate #3 Sensitivity (S)
RN_SENS NR Lite2 Sensitivity
LI_200R_MULT LI-200R Multiplier
LI_190R_MULT LI-190R Multiplier

After entering the required information, save the configuration file to your computer.

Entering sensor calibration information (Standard Package)

Each program includes some sensors that have specific calibration information. In Blueprint Utility, click the Configuration tab. The required information is described in a note in the program. In this collection of sensors, the soil heat flux plates, NR01 net radiometer, and LI-190R require calibration information.

Table 4‑2. Constants and the corresponding calibration values that must be entered to use the Standard Biomet Package.
Constant Value to enter
SHF_SENS_1 Soil Heat Flux Plate #1 Sensitivity (S)
SHF_SENS_2 Soil Heat Flux Plate #2 Sensitivity (S)
SHF_SENS_3 Soil Heat Flux Plate #3 Sensitivity (S)
SWIN_SENS NR01 Short Wave Upper Sensor Sensitivity
SWOUT_SENS NR01 Short Wave Lower Sensor Sensitivity
LWIN_SENS NR01 Long Wave Upper Sensor Sensitivity
LWOUT_SENS NR01 Long Wave Lower Sensor Sensitivity
LI_190R_MULT LI-190R Multiplier

After entering the required information, save the configuration file to your computer.

Entering sensor calibration information (Premium Package)

Each program includes some sensors that have specific calibration information. In Blueprint Utility, click the Configuration tab. The required information is described in a note in the program. In this collection of sensors, the soil heat flux plates, CNR4 net radiometer, and LI-190R require calibration information.

Table 4‑3. Constants and the corresponding calibration values that must be entered to use the Premium Biomet Package.
Constant Value to enter
SHF_SENS_1 Self-Calibrating Soil Heat Flux Plate #1 Sensitivity (Sreference)
SHF_SENS_2 Self-Calibrating Soil Heat Flux Plate #2 Sensitivity (Sreference)
SHF_SENS_3 Self-Calibrating Soil Heat Flux Plate #3 Sensitivity (Sreference)
R_SELF_1 SC Soil Heat Flux Plate #1 Resistance (Rheater)
R_SELF_2 SC Soil Heat Flux Plate #2 Resistance (Rheater)
R_SELF_3 SC Soil Heat Flux Plate #3 Resistance (Rheater)
SWIN_SENS CNR4 Short Wave Upper Sensor Sensitivity
SWOUT_SENS CNR4 Short Wave Lower Sensor Sensitivity
LWIN_SENS CNR4 Long Wave Upper Sensor Sensitivity
LWOUT_SENS CNR4 Long Wave Lower Sensor Sensitivity
LI_190R_MULT LI-190R Multiplier

After entering the required information, save the configuration file to your computer.

Sending the program

Click Push to send the program. The parameters will take effect immediately.

Additional settings must be applied in order to log biomet data in eddy covariance datasets. These are described in the following section.

Connecting the biomet system to the EC system

When the system is properly configured, biomet data will be logged in each .ghg data file, which is processed by EddyPro on the SmartFlux System. The data acquisition system is compatible with the LI-7500DS (and SmartFlux 3 System) and the LI-7500A/RS or LI-7200/RS (and SmartFlux 2 System).

  1. Connect data cables for the system and power on the components.
  2. The DAqM, DRM, SmartFlux System, network switch, and gas analyzer all must be powered on.
  3. Launch the gas analyzer software and connect to the gas analyzer.
  4. The software is called something like LI-7x00 A RS DS x.x.x.
  5. Establish communication between the gas analyzer with the data acquisition system.
  6. Click the Biomet button. Under Datalogger type, select Data Acquisition System (DAqM). Select the SmartFlux System, click Connect, then click OK.
  7. Open biomet dialogSet data logger type in win-GHG
  8. View live data.
  9. Under the Data tab, you'll see live data values on the display. Readings are updated about once per minute.
Live data view in the win-GHG app

Configuring the PhenoCam

Note: The StarDot NetCam SC (PhenoCam) has been discontinued by the manufacturer and is no longer available for purchase.

The PhenoCam network1 provides extensive documentation for using this model of camera. Refer to their documentation (https://phenocam.nau.edu/webcam/) to configure the device according to those standards. Here, we provide basic instructions for configuring the device for LI-COR eddy covariance systems.

Setting the PhenoCam IP address

Download the StarDot Tools application from http://www.stardot.com/downloads and install it on your Windows computer.

  1. With the network and power cables connected as described earlier, power on system, including the camera.
  2. Launch the StarDot Tools application.
  3. The application will present all of the PhenoCams on your network. Identify the one you want to configure and record its ID and IP address.
  4. Enter the IP Address into a web browser address bar.
  5. The browser will display a recent image from the camera as well as some menu links.
  6. Click Config in the lower left of the page and log in when prompted.
    • User Name: admin
    • Password: admin
  7. Configure the image capture/upload timing.
  8. Click the Network tab and configure the settings:
    1. Under IP Assignment, select Manual.
    2. IP Address: 192.168.13.205 if following LI-COR recommendations (see ).
    3. Set the Subnet Mask: 255.255.255.0, typically.
  9. If you are using a cellular or satellite gateway to access your system remotely, under Network, set the IP Address to that of the cellular or satellite gateway. This will be provided by your service provider.
Table 4‑4. Recommended IP addresses for remote instruments that include the LI-7500DS and SmartFlux 3 System.
Device Public Start Port Public End Port Protocol Host IP Private Start Port
SmartFlux 3 7200 0 TCP 192.168.13.200 7200
SmartFlux 3 7222 0 TCP 192.168.13.200 22
SmartFlux 3 5050 0 TCP 192.168.13.200 5050
SmartFlux 3 50111 0 TCP 192.168.13.200 50111
SmartFlux 3 50112 0 TCP 192.168.13.200 50112
SmartFlux 3 50311 0 TCP 192.168.13.200 50311
LI-7700 7700 0 TCP 192.168.13.202 7700
Phenocam 7980 0 TCP 192.168.13.203 80
Sutron 2300 0 TCP 192.168.13.204 23
Campbell 6785 0 TCP 192.168.13.205 6785
LI-8100A 1526 0 TCP 192.168.13.206 1526
LI-8250 8200 0 TCP 192.168.13.209 80
LI-8250 8201 0 TCP 192.168.13.209 8080
LI-8250 8202 0 TCP 192.168.13.209 22
LI-7810 7810 0 TCP 192.168.13.210 80
LI-7810 7811 0 TCP 192.168.13.210 8080
LI-7810 7812 0 TCP 192.168.13.210 22
LI-7815 7815 0 TCP 192.168.13.211 80
LI-7815 7816 0 TCP 192.168.13.211 8080
LI-7815 7817 0 TCP 192.168.13.211 22
LI-7820 7820 0 TCP 192.168.13.212 80
LI-7820 7821 0 TCP 192.168.13.212 8080
LI-7820 7822 0 TCP 192.168.13.212 22
LI-7825 7825 0 TCP 192.168.13.213 80
LI-7825 7826 0 TCP 192.168.13.213 8080
LI-7825 7827 0 TCP 192.168.13.213 22

If your camera is on an unsecured network, it may be vulnerable to hacking. Therefore, we discourage using the camera in unsecured network environments.

Connecting the PhenoCam to the EC System

In the LI-7500A/RS/DS or LI-7200/RS software, connect with the gas analyzer that will be part of your eddy covariance system.

  1. Connect the camera network cable to the EC system network switch.
  2. Power on the system and the camera.
  3. Be sure that the camera is powered with 12 VDC.
  4. Click the SmartFlux button and select the PhenoCam tab.
  5. Check Connect to PhenoCam.
  6. Select StarDot Camera from the menu.
  7. Enter the Hostname or IP Address (192.168.13.205 recommended) of the camera in the HostName field.
  8. Click Apply Camera Settings.

In this configuration, the system will store one image to the USB drive each day until about 7% of the capacity is occupied by images, at which time the old images will be deleted to make space for the new ones. If you use the FluxSuite server, your images will be posted once daily.

Turning ON datalogging

As part of an eddy covariance system, the DAqM and DRM provide data to the SmartFlux System for logging in eddy covariance data files. Logging is controlled in the LI-7500A/RS/DS or LI-7200/RS software. In the software dashboard, click Start or Stop to control logging. The logging setting will apply immediately.

The win-GHG software interface presents a dashboard that indicates data logging state and presents controls

All data files are stored on the USB drive on the SmartFlux System. Most files are saved on the USB drive in the LI-7550, except for the data acquisition system backup file. On the SmartFlux USB drive, the data are in the compressed .ghg files and a separate folder of biomet data backup files (called something like licor/data/daqm/YYYY-MM-DD-HHMMSS-daqm.zip). See Viewing and evaluating data for more details.